    Markus Stoeckli
    Moderator

    Yes, it works. We use it in our lab and it’s relatively robust and reproducible. In a typical coating we spray 8 ml of saturated CHCA solution in 50:50 ACN/0.1%TFA using a TLC sprayer. The authors state:[quote] The sample is allowed to dry for 1-2 min before the next coating cycle[/quote]
    IMHO it is a requirement to never let the sample completely dry in between spray cycles ❗ .

    Markus

    #840
    sclerens
    Participant

    Markus, could you elaborate on the flow rate during spraying, and what kind of surface is covered by the 8 mL of matrix solution?
    Currently we use a home-built nebulizer, but only spray at 40?L/min because at higher flow rates the sections get too wet and the liquid is blown laterally over the sections due to the gas flow.

    Stefan

    #842
    Markus Stoeckli
    Moderator

    I can only estimate the flow rate of of our spray since we use a TLC sprayer operated by Bernoulli’s principle. We ajdust the spray to form a fine mist setting the pressure to approx 0.8 bar. This results in an estimated flow of 1 ml/min. We spray 8 ml in 30 minutes, with cycles of 10 s spray and 30 s pause.

    Markus

    #855
    Diesel
    Moderator

    Additionally I want to remark that we are using 8ml for coating 4 MALDI plates (40x40mm) at the same time. So for one of these plates approx. 2ml.

    Dieter