  • #909
    Markus Stoeckli
    Moderator

    We are typically not too concerned about the damage to the microstructure during the drying process since we image with a maximal resolution of 100 micron. Freeze-drying is commonly used in histology to preserve the tissue structure, but also here we observe some tissue shrinkage.

    markus

    #911
    Nathan Jones
    Participant

    I would like to try to get resolution into the 20-40 um range. The raster on the 4700 goes down to 10um i believe. Do you typicaly fix standard “spots” of whatever analyte you are looking for? Recently I started adding spots next to the sections with a std prep of the neuropeptide of interest.

    #913
    Markus Stoeckli
    Moderator

    The accuracy of the sample stage is just one thing to consider, but equally important are laser spot diameter and crystal size (see below). Also, smearing of analytes during wet matrix deposition will decrease spatial resolution. We use standards on the same plate to control the reproducibility of the matrix coating and for calibration.

    markus

    [img]http://www.maldi-msi.org/images/forum/DHB.jpg[/img]
    Spray of DB in 70:30 methanol/water

    [img]http://www.maldi-msi.org/images/forum/SA.jpg[/img]
    Spray of SA in 50:50 ACN/water