  • in reply to: LEAP Technologies TM Sprayer #3126
    Alain CREISSEN
    Participant

    Dear Milad, Rob and Junhai

    Great conversation. I just started being active in this forum and noticed your conversation. We can certainly help you with these questions if they are still of interest.

    I agree with Rob that 0.200ml/min can be too much flow and results in a very wet spray, especially with 50% H20.

    My starting point with CHCA sprayed on tissues, is:

    <Solvent>70% ACN + 0.1%TFA</Solvent>
    <Temp>75</Temp>
    <NumPasses>4</NumPasses>
    <Concentration>10</Concentration>
    <FlowRate>0.12</FlowRate>
    <Velocity>1200</Velocity>
    <TrackSpacing>3</TrackSpacing>
    <Pattern>HH</Pattern>
    <Pressure>10</Pressure>
    <GasFlowRate>3</GasFlowRate>
    <DryTime>0</DryTime>
    <NozzleHeight>40</NozzleHeight>

    Note that this protocol has 70% ACN (instead of 50%). Higher organic content makes the spray evaporate faster. The temperature of 75 Deg C is close to the boiling point of the mixture but should be stable and will evaporate quickly.

    Feel free to email me at acreissen(at)htximaging.com if you need more information

    Alain