  • in reply to: Sample drying #913
    Markus Stoeckli
    Moderator

    The accuracy of the sample stage is just one thing to consider, but equally important are laser spot diameter and crystal size (see below). Also, smearing of analytes during wet matrix deposition will decrease spatial resolution. We use standards on the same plate to control the reproducibility of the matrix coating and for calibration.

    markus

    [img]http://www.maldi-msi.org/images/forum/DHB.jpg[/img]
    Spray of DB in 70:30 methanol/water

    [img]http://www.maldi-msi.org/images/forum/SA.jpg[/img]
    Spray of SA in 50:50 ACN/water

    in reply to: Sample drying #909
    Markus Stoeckli
    Moderator

    We are typically not too concerned about the damage to the microstructure during the drying process since we image with a maximal resolution of 100 micron. Freeze-drying is commonly used in histology to preserve the tissue structure, but also here we observe some tissue shrinkage.

    markus

    in reply to: large files in BioMAP #905
    Markus Stoeckli
    Moderator

    At least twice the size of the image file.

    Markus

    in reply to: gamma variate #902
    Markus Stoeckli
    Moderator

    Please refer to page 52 of the BioMap manual included with BioMap version 3.6.6. Let us know if that doesn’t help…

    in reply to: 4700 stucked #896
    Markus Stoeckli
    Moderator

    The behavior you are describing lets me assume that you are trying to acquire an image with the ‘Acquisition Mode’ in the 4700 acquisition method set to ‘automatic’. That works only if you defined a spotset matching your image raster. Otherwise, the instrument will always acquire a spectrum from the spot nearest to each image point.

    markus

    in reply to: BIOMAP Help #894
    Markus Stoeckli
    Moderator

    This short movie may get you started: [url=http://www.maldi-msi.org/download/BioMap-Browsing.wmv]BioMap-Browsing.wmv[/url]

    Post edited by: net-climber, at: 2007/03/21 19:36

    Post edited by: net-climber, at: 2007/03/21 19:36

    in reply to: Biomap license problem #892
    Markus Stoeckli
    Moderator

    That looks odd. Let me try to narrow things down…

    The license generator works with the [b]MAC address[/b] of the first network interface in your computer. We had problems before when users would enable/disable a WLAN interface and the software would get a different MAC address. To see your current number, type ‘ipconfig /all’ in a DOS window.

    Make sure that your licenserc file does not have an [b]extension[/b]. In Windows, the extension ‘.txt’ may be hidden. To overcome this problem, select the menu ‘Tools – Folder Option’ in the file manager. Switch to the ‘View’ tab and in there disable the option ‘Hide extensions for know file types’.

    The [b]location[/b] of the licenserc file is critical. Simply copy it into the ‘..\BioMap\settings’ folder and execute ‘Install settings.bat’ located in the BioMap folder.

    Hope that help – if not, please let us know…

    markus

    in reply to: Importing a MSI file #890
    Markus Stoeckli
    Moderator

    Yes, we had that in our lab before. Let me guess: Your mass spectra contain more than 32768 data points. This is a known limitation of BioMap which works internally with 16-bit signed integers.

    markus

    in reply to: Error in function PBMT_PLOT_UI_EVENT EXECUTE not #888
    Markus Stoeckli
    Moderator

    Thanks for reporting that bug. We’ll take care of it and it will be fixed in the upcoming release.

    markus

    in reply to: Overlay #884
    Markus Stoeckli
    Moderator

    To open a picture (e.g. jpg or tif): File-Import-Pixmap

    To copy a ROI: Edit-Copy, Edit-Paste-ROI

    markus

    in reply to: Exporting spectra for Data Explorer #882
    Markus Stoeckli
    Moderator

    While there is no direct way to save files in DE format, they may be imported as ASCII. Here’s how:
    [ul]Save spectra in BioMap as ‘List’. A text file will be created with a name similar to ‘your_file_plot.txt’.[/ul]
    [ul]Localize this file and open it with a text editor. The content has this format:
    [code:1] X L1
    498.302 89.0408
    498.555 87.3492
    498.807 85.3810
    499.059 85.1474
    499.311 89.4717
    499.563 88.5828
    499.815 86.0317[/code:1][/ul]
    [ul]Delete the first line and replace it with the following two lines:
    [code:1]TITLE "Your Title"
    TYPE MASSSPEC[/code:1][/ul]
    [ul]Save the file.[/ul]
    [ul]Open DE and open a random spectrum.[/ul]
    [ul]Select the menu item File – Import – ASCII Spectrum…[/ul]
    [ul]Localize the text file and load it.[/ul]

    markus

    in reply to: Error in function PBMT_PLOT_UI_EVENT EXECUTE not #881
    Markus Stoeckli
    Moderator

    This error occurs if you run BioMap version 3.6.2 or older on the IDL Virtual Machine. With BioMap version 3.6.3 this limitation was overcome and spectra may be exported with any IDL version.

    markus

    in reply to: Problems using MMSIT #878
    Markus Stoeckli
    Moderator

    I checked it out on our sSTR and found nothing wrong with the MMSIT. For the mass range 2000-25000 with bin size 4 ns we end up with 40000 data points in reflector mode and 27000 in linear mode. 💡 Could you verify for me that the data file which you specified in the MMSIT was acquired with the appropriate number of data points (open it as instrument settings in the VC)? It is this file which defines the image settings, no the current settings of the VC.

    Markus

    in reply to: What is a voxel? #876
    Markus Stoeckli
    Moderator

    voxel = vo(lume) + (pi)xel

    Therefore we talk of pixels in 2d applications and voxels in 3d tools like BioMap.

    in reply to: Data Explorer v 4.??? #874
    Markus Stoeckli
    Moderator

    The sofware ‘DataExplorer’ you are looking for is a product from Applied Biosystems. Contact them directly (http://www.appliedbiosystems.com) for more information.

    markus